error code 222.4180

What can cause this error code?

   RE: Infusion Pump Report Post
Wed May 05 2021Reply from rsiefers
avatar placemark

either the door or air in line assembly


Report Post
Wed May 05 2021Reply from T-Lobato
T-Lobato

Hi Long,

From the manual:

Subsystem: Safety Processor Cross-Check

Explanation: Improper door and/or flow stop state detected during infusion.

Repsonse: 1. Replace AIL Assembly; 2. Replace Door Assembly

T.

 


TJ Lobato | MedWrench Guru