Error on monitor

'Water in Chamber' error appeared on Philips monitor. Thought it was the gas module since I'm not sure where else the water could be coming from. I replaced the module, but could anyone give clarification on this error?

   RE: Medical Gas Monitor Report Post
Sat May 08 2021Reply from CRC Biomedical Services
CRC Biomedical Services

Replace the water trap. If that doesn't work you can probably fix it by just performing a PM. This is a GE device, not Philips. 

Please be advised that because the fix to this is so basic, yet these questions had to be asked - make sure you are qualified to work on, and verify the operation of medical equipment to the manufacturer's standards. This is not something I would expect an entry level technician with any amount of training to have to ask. Where does the water come from in a device with a water trap??? Hmmm... 


Mickey A. Couvertier, CBET, CLRT
President & Senior Technician
CRC Biomedical Services