B51 error message
RE: Electrosurgical Unit (ESU) Report Post
Domlaco2,
Do you still get the error if all accessories are unplugged (i.e, footswitches, etc.)? If so, maybe it's a bad accessory.
If that's not the problem, I would perform an EST to verify there is not a short and to make sure the ground is intact. If those pass, inspect the footswitch connectors in the back of the unit.
Still nothing? You may have to send it in.
Just me.
T.
TJ Lobato | MedWrench Guru
Report Post
Hello T,
Thank you so much for your help!
This problem comes when I plug the footswitch on the unit, I have the message "footswitch has not been assignated to a socket", and after that, I have the 2 other error messages (B03 and B51). Anyway, thank again for your time, appreciated!
Best regards
Report Post
Remove the IIF module for socket 2 and try if the error persists,or change tje modules between each other.
Report Post
How are your sockets configured? From top to bottom?
Bipolar/monopolar/monopolar/neutral?
Report Post
Hello,
I took several pictures of the unit and error messages, can I please send them to you?
Thanks!