CaseSaver Feature Activated
The case saver feature does not turn off. The PowerSuites have 4 heads internally that together make 100W of holmium laser. When 1 or more of the heads is not functioning efficiently enough, the laser runs at limited power. If the max is 60W @ 40Hz, then 1 head is down. If 30W, then 2 heads are down. Open up the laser, and check the rods and mirrors.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Any kind of optical damage. They will appear as spots/burns. Typically, in order to cause it to go into case saver, the spots will be greater than or equal to 1 mm in diameter.
Also, the laser's burns should be round 4 mm spots. If they are slitted like cat's eyes, that could also be an issue. If that is the case, then alignment is required.
Wayne Delgado
www.umlasers.com
Hi Wayne, thanks again.
By any chance is ther any way you can send me step by step instructions or maybe a video on how to check and fix the issue please. I don't want to make it worse.
Joe..
Joe,
Please call the office. We can speak more over the phone.
Wayne Delgado
www.umlasers.com
Hey Wayne,
if you can please check out the youtube video below of the laser being powered up, maybe this can give you an idea of what might be causing the problem. Let me know what you think and I will drop you a call tomorrow.
https://www.youtube.com/watch?v=-2LcP1MiPz8&ab_channel=jtridexter
Joe..