Never ending start up screen, on a 740i
RE: Lensometers Report Post
Antenakias,
The service manuals are attached to this page on the left side. I glanced through them. My takeaways: check the hand switch and/ or reload the software. To me it sound like it is not passing POST, power on self test, which could be voltages or memory or something else completely. I also saw that this unit can be connected to the network but I don't think a broken connection would not let it boot up.
Just me,
T.
TJ Lobato | MedWrench Guru
Report Post
Happy news, i managed to fix the problem by resoldering the main cpu board. And i was lucky .
Now its more than two months that it is working.
Have good holidays
Report Post
HFA 740i
Happy news, i managed to fix the problem by resoldering the main cpu board. And i was lucky .
Now its more than two months that it is working.
Have good holidays
Report Post
HFA 740i
Boots with the message below
System startup state 19
Aperture may be misaligned
(Pressing Proceed will NOT allow testing)
[Retry] [Proceed]
Even after pressing [Retry] it persists. Same as when I restart the machine.
Why this message?
What can I do about it?
Please help out.
Thank you.