Check Vent: Proximal Pressure Sensor Auto Zero Failed Error

Check Vent: Proximal Pressure Sensor Auto Zero Failed Error.Any Suggestions ?

   RE: Ventilator Report Post
Mon Jun 07 2021Reply from Ben C
Ben C

One of these companies may be able to assist:

Agiliti  

Mobile Medical


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Tue Jun 08 2021Reply from Chris S
Chris S

1. Verify pin alignment between solenoids and the DA PCBA.

2. Replace the proximal auto-zero solenoid (SOL 3 and SOL 4).

3. Replace the DA PCBA.