Tue Jun 08 2021

Light not coming on.

I have an Olympus microscope model BX51TRF the light is not coming on and when i measure the voltage it's only showing 2v that's reaching the terminals instead of 6v. Can you help please.

   RE: Microscopy Report Post
Thu Jul 07 2022Reply from nfocus_7061
avatar placemark

Does the 2V vary at all when you move the Rheostat?  I have had a number of control boards go out on these scopes.  You might consider also a Nanodyne Conversion Kit. 

 

https://www.nano-dyne.com/led-microscope-lights/olympus-bx51-bottom-illuminator/


Jim Janoso
Northern Focus Optical Inc.
Sales and Service Manager

Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.