Is anybody connected this analyse to LIS via RS232 ? If yes, please tell me the cable pin out computer DB9 and analayzer DB9
Wed Jun 16 2021
LIS Communication
Wed Jun 16 2021Reply from Ben C
What is the name of manufacturer and model of the equipment you are working on?
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Wed Jun 16 2021Reply from sergiu.vilcu_8892
Thu Jun 17 2021Reply from Amscopolaris
Pin out
pin 1... sample ground... GND
pin 2...RS232IN ... . RxD
pin 3 ... RS232OUT... . TxD
pin 4...NC
pin 5... Digital ground... DGND
pin 6...NC
pin 7.. RTS
pin 8...CTS
pin 9...Print VCC
(NC=Not Connected)
Report Post
Thu Jun 17 2021Reply from sergiu.vilcu_8892
Thu Jun 17 2021Reply from Amscopolaris
Thu Jun 17 2021Reply from sergiu.vilcu_8892
Also have try diferent methode of BaudRate, stop bit..parity...nothing..think is the cable issue.