Wed Jun 23 2021

Not charging

Hello, good morning we just get an amx the machine was working really fine, but after a while the batteries get dead, we tried to charge it back but it does not charge the display board dont show anything we tried to turn switch on to moe and it does get any charge. thank you, hope you help me

   RE: Mobile Radiography Report Post
Wed Jun 23 2021Reply from Ben C
Ben C

What is the name of the manufacturer and model of the equipment you are working on? 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Wed Jun 23 2021Reply from dann14
avatar placemark

Name of manufacturer: general electric compnay, 

model: 46-270157G1 

thanks for your help


Report Post
Wed Jun 23 2021Reply from ffriedmann
avatar placemark

What is the voltage from the batteries. If less than 90 (?) where the LVPS will drop out, then you will have to test/charge each battery individually. Very possible since you have no displays on the unit. The service docs (class A only of course) are available on the GE site 

https://customer-doc.cloud.gehealthcare.com/#/cdp/dashboard

Review the schematic pages for the charging function, but remember with no LVPS output, no CPU, no charger board control, no charging possible. 


Report Post
Mon Mar 13 2023Reply from www.PI.solutions
www.PI.solutions
Know this is an old thread but here goes: Put new batteries in but they died after a week. Took batteries home and charged them up. Replaced charger bd, CPU, rotor bd Ran volt and charger. Ask. Force charging isn’t working. Any ideas?


Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions