Is there a calibration procedure for a Detecto 6735 Infant Scale?
Thu Jun 24 2021
How to calibrate?
Fri Jun 25 2021Reply from Amscopolaris
With the display showing CAL
press the PRINT key. The display will change to show the
current setting 0 (0=NO). If the scale has been previously calibrated and you wish to skip
calibration and proceed to A9CC
, the Acceleration of Gravity, simply press the PRINT key and
the internal calibration factor will be retained.
To begin calibration, press UNITS to select 1 (1=YES), then press the PRINT key. After
pressing the PRINT key the display will change to LOAd
.
Report Post