I have replaced the garments and the tubing, and I have replaced the valve manifold. Equipment still won't pass T6. I get the low pressure error. Any ideas?
Wed Jul 07 2021
SCD Express failing Test 6 - Leak Test
Wed Jul 07 2021Reply from Jeremy.Kirlin
Wed Jul 07 2021Reply from RosaAP
I have not. Funny thing is that T2, T3, T4, T5, T7 all passed and equipment operates well during clinical operations. There are no low pressure errors during clinical operations either.
Report Post
Wed Jul 07 2021Reply from Chris S
The only other thing besides the maint board it could be besides the maint board is the compressor or associated tubing. Do you have a spare unit you could swap parts from?