I have a strange issue.
I have 2 units . 1 unit has defective lamp, I tried replacing with a new lamp module the lamp does not work., but when i took the lamp module from other unit and install in the first unit it start to work,
So this clearly shows that Power supply if working as well.
Please help me in identifying the problem and recitfing it
Thu Jul 08 2021
New Bulb Not Fire
Thu Jul 08 2021Reply from T-Lobato
Have you tried taking the new module and placing it in the second unit? If it doesn't work, it's probably a bad module.
Just me.
T.
TJ Lobato | MedWrench Guru
Report Post
Thu Jul 08 2021Reply from Chris-CBET
Mon Jul 19 2021Reply from Better Biomed Channel
Is the light source clicking like it's trying to ignite the xenon bulb? The clicking means that the HV part of your power supply is working. If it's not clicking, do the fans spin? There is a balance between the bulb, the conductive heatsinks and the sensitivity of the HV switchover (in the PSU).
Report Post
Mon Jul 26 2021Reply from Riz
Fri Mar 25 2022Reply from Bryan8796
Build is flickering on my mlx300. Turns on for a fraction of a second every 4 seconds. Any idea what is causing the issue