New Bulb Not Fire

I have a strange issue. I have 2 units . 1 unit has defective lamp, I tried replacing with a new lamp module the lamp does not work., but when i took the lamp module from other unit and install in the first unit it start to work,  So this clearly shows that Power supply if working as well. Please help me in identifying the problem and recitfing it

   RE: Surgical Light Report Post
Thu Jul 08 2021Reply from T-Lobato
T-Lobato

Have you tried taking the new module and placing it in the second unit?  If it doesn't work, it's probably a bad module.

Just me.

T.


TJ Lobato | MedWrench Guru
Report Post
Thu Jul 08 2021Reply from Chris-CBET
Chris-CBET

you have a bad module, check your contacts, check your cable for continuity


Report Post
Mon Jul 19 2021Reply from Better Biomed Channel
Better Biomed Channel

Is the light source clicking like it's trying to ignite the xenon bulb? The clicking means that the HV part of your power supply is working. If it's not clicking, do the fans spin?  There is a balance between the bulb, the conductive heatsinks and the sensitivity of the HV switchover (in the PSU).  


Report Post
Mon Jul 26 2021Reply from Riz
avatar placemark

It was bad module.problem solved by replacing with another one 


Report Post
Fri Mar 25 2022Reply from Bryan8796
avatar placemark

Build is flickering on my mlx300. Turns on for a fraction of a second every 4 seconds. Any idea what is causing the issue