Fri Jul 09 2021

Constant Error N251

Hospira/ Abbott Plum A+ is giving constant error N251. The service manual addresses this error code and the listed corrective action is to "Back- Prime the test set or Replace the test set". Well, I'm here to tell you that even after replacing the test set, the mechanism assembly, the main PCB and back- priming, the pump is still throwing a constant error N251. Do you have any idea what else could cause this or how to correct the issue?

   RE: Infusion Pump Report Post
Fri Jul 16 2021Reply from Ben C
Ben C

One of these companies may be able to provide a solution:

Elite Biomedical 

Mobile Medical Repair 

iMed 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Mon Jul 19 2021Reply from IVSupport
avatar placemark

The error is 1 of 2 things. Bad fluid shield or bad mechanism. I'd replace the fluid shield first. If it doesnt work that means on the mechanism the top potentiometer is bad. Replace the mechanism


Report Post