Scan stops while doing patient and operator has to completely restart the equipment to continue.
We have the errors 11-10F0-02, 11-1010-02,11-10F2-02
Wed Jul 14 2021
Scan stops and operator has to restart the equipment to continue
Wed Jul 14 2021Reply from Ben C
What is the name of the manufacturer and model of the equipment you are working on?
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Wed Jul 14 2021Reply from J4RS20
Thu Jul 15 2021Reply from armonj_3859
First check OGP connections, and SlipRing brushes. the next step might be OGP to KVP cable, KVP bd. or OGP bd.
Report Post
Thu Jul 15 2021Reply from J4RS20
We cleaned the slip ring already and verified that the brushes are Ok. The problem is intermittent.
Report Post