Wed Jul 14 2021

Scan stops and operator has to restart the equipment to continue

Scan stops while doing patient and operator has to completely restart the equipment to continue. We have the errors 11-10F0-02, 11-1010-02,11-10F2-02

   RE: CT Scanner Report Post
Wed Jul 14 2021Reply from Ben C
Ben C

What is the name of the manufacturer and model of the equipment you are working on? 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Wed Jul 14 2021Reply from J4RS20
avatar placemark

We are working on a GE HiSpeed Dual Series


Report Post
Thu Jul 15 2021Reply from armonj_3859
avatar placemark

First check OGP connections, and SlipRing brushes. the next step might be OGP to KVP cable, KVP bd. or OGP bd.


Report Post
Thu Jul 15 2021Reply from J4RS20
avatar placemark

We cleaned the slip ring already and verified that the brushes are Ok. The problem is intermittent. 


Report Post