We have a couple of M10 Concentrators and we are trying to test the output occlusion alarm without any success. There doesn't seem to be any internal leak and the pressures all are passing. Anyone have any ideas?
Wed Jul 14 2021
Occlusion Alarm not working
Fri Jul 16 2021Reply from CarlAnderson
We discovered that the outlet connector had multiple small cracks and this was causing leaks. The original connector was plastic, the replacement connectors are metal, so i suspect this is a known issue.