Problem with steam boiler 433HC
RE: Sterilizers/Autoclaves Report Post
Try disconnecting the leak detection probe from the board and see if it clears the issue.
Report Post
Hi Chris, thank you for responding me... the leak detection probe, is that the service manual calls spill probe? The one that is connected to the pan, or is the lowest probe wich is connected to the little board? The steam boiler is a CAS 30 (the one with the vertical water deposit)
Thanks
Report Post
Yes, the short probe mounted just above the drip pan. It is there to detect a leak and shut down to prevent flooding. I've had them short in the wiring or have a piece of the foil insulation ground out the probe and cause the fault light.
Report Post
Hi Chris and everyone who is reading, I disconnected the leak probe, and the fault continues, but, when I disconnect the Hi probe (which is connected to the same connector in the board with the leak probe) the fault is cleared and the boiler works. However, with the hi probe disconnected, the boiler will not be able to control the filling, right? So, what do you think?