service manual

WHAT IS THE REMEDY FOR ERROR 195

   RE: Electrosurgical Unit (ESU) Report Post
Wed Jul 28 2021Reply from Chris S
Chris S

Internal diagnostics. Recall button may be stuck (button shown on terminal using serial port)

1. Turn off, then turn on the generator. Do not press buttons or accessory activation devices during the self-test.

2. If the alarm number reappears, disconnect all accessories. Turn off, then turn on the generator again.

3.  Replace the Keypad

If the alarm number reappears, record the number and call the Covidien Service Center.