Fri Jul 30 2021

Bluescreen error code 2726

I deveped a blue screen with error 2726 after using an Hfl 38x. on a Sonosite M Turbo. I have since not been able to use any other probe. I tried resetting the system by depressing "0" when switching on but have not been able to achieve a reset. Pls can someone help? Whizkid

   RE: Ultrasound Systems Report Post
Wed Oct 06 2021Reply from Gordjan
avatar placemark

Hello, did you fixed that problem, do you have any suggestion, I have the same problem.

Thanks,

Goran


Report Post
Sat May 21 2022Reply from Alex22
avatar placemark

Please let us know how uou fixed the proble. Many users get the same issue 2726 error. Thanks


Report Post
Mon May 23 2022Reply from Bekki_E
avatar placemark

Hi Slex,

 

The only fix for an Assert 2726 is a main board replacement. This assert code is caused by attaching a faulty transducer to the system that causes damage to the main PCBA. Please call Sonosite tech support for a quote for repair.