System Failure (907)
RE: Pulse Oximeter Report Post
I think that refers to an RTC error (real time clock). I copied these from another answer. It might help you get in to reset the clock.
try passcodes
Static passcodes
Covidien Master Code 1287
Service Mode Password setting 6290
Covidien Service Password Setting 1309
DEFAULT PASSCODES
Homecare Password setting 1111
Homecare/Sleep to Standard Password setting 2222
Sleep study Password Setting 1212
Todd G. Schmechel
Report Post
Error 907 is real time clock, internal battery is dead. Not the main battery kind of like a CMOS battery
Report Post
It's usually the coin cell battery that needs to be replaced. There is a little compartment under the batteries. Some editions of the PM10N do not have a CC battery in a small compartment. With those models it's located on the main board, which you will need to remove to access.
Report Post
When you change the CR1220, you need to enter in the service menu and clear all errors then reboot the equipment and the error no appears.