Wed Aug 04 2021

System Failure (907)

The equipment turns on and will sporadically read but it continues to flash with System Failure (907) on top.

   RE: Pulse Oximeter Report Post
Thu Aug 05 2021Reply from Todd G. Schmechel
Todd G. Schmechel

I think that refers to an RTC error (real time clock).  I copied these from another answer.  It might help you get in to reset the clock.

try passcodes

Static passcodes

Covidien Master Code                          1287
Service Mode Password setting            6290
Covidien Service Password Setting      1309

DEFAULT     PASSCODES    

Homecare Password setting                                  1111
Homecare/Sleep to Standard Password setting    2222
Sleep study Password Setting                               1212


Todd G. Schmechel

Report Post
Tue Dec 07 2021Reply from Mongai Gussin
avatar placemark

Problem still persists 


Report Post
Tue Dec 07 2021Reply from Mongai Gussin
avatar placemark

Error 907 still persists 


Report Post
Tue Jan 25 2022Reply from james jessee
avatar placemark

Error 907  is real time clock, internal battery is dead. Not the main battery kind of like a CMOS battery


Report Post
Wed Jun 28 2023Reply from Paul M
avatar placemark

It's usually the coin cell battery that needs to be replaced. There is a little compartment under the batteries. Some editions of the PM10N do not have a CC battery in a small compartment. With those models it's located on the main board, which you will need to remove to access.


Report Post
Thu Feb 15 2024Reply from celarrieta_5854
avatar placemark

When you change the CR1220, you need to enter in the service menu and clear all errors then reboot the equipment and the error no appears.


Report Post
Wed Nov 13 2024Reply from Carlos De Sousa
Carlos De Sousa

Thank you guys for the help! 