Erorr 24

Erorr 24 Siemens - Mobilett XP Hybrid

   RE: Mobile Radiography Report Post
Thu Aug 05 2021Reply from Tomelgy
avatar placemark

Error 24 is the Anode did not rotate.

Thinking simply: This could be a failed cable within the arm to tube, at the tube rotation point (nuckle) or could be a seazed X-Ray tube anode, or tube stator winding fault.

Obviously it could be the supply fuse to the rotation circuit. Check D927 F5, F6.

Sometimes when the tube collimator is rotated, it damages the wires becasue the end stops don't work.

Tom