Hello Guys,
I have a Valleylab force2 which works well with the Valleylab Non-Rem Patient Return Electrode (REF: E7506), but deosn't work with the Valleylab Rem Patient Return Electrode (REF: E7507).
Please I need help to make it work with the Valleylab Rem Patient Return Electrode (REF: E7507).
Thanks in advance.
Sat Aug 07 2021
Patient Return Electrode
Mon Aug 09 2021Reply from Amscopolaris
Wed Aug 11 2021Reply from Denova Medical
Do a REM Contact Quality System Test per the service manual. If that test fails, you may need a calibration or you may have something broken. A common failure is the microswitch which is actuated by the center plastic pin in the REM accessory plug. Based on your discription, there is a good chance your failure is that switch. Calibration procedure for the REM is in the service manual.
Photios Peter Dalamagas Denova Medical Inc (704)-233-7574