Patient Return Electrode

Hello Guys, I have a Valleylab force2 which works well with the Valleylab Non-Rem Patient Return Electrode (REF: E7506), but deosn't work with the Valleylab Rem Patient Return Electrode (REF: E7507). Please I need help to make it work with the Valleylab Rem Patient Return Electrode (REF: E7507). Thanks in advance.

   RE: Electrosurgical Unit (ESU) Report Post
Mon Aug 09 2021Reply from Amscopolaris
Amscopolaris

replace the Return Electrode connector with a new generation type connector ,


Report Post
Wed Aug 11 2021Reply from Denova Medical
Denova Medical

Do a REM Contact Quality System Test per the service manual. If that test fails, you may need a calibration or you may have something broken. A common failure is the microswitch which is actuated by the center plastic pin in the REM accessory plug. Based on your discription, there is a good chance your failure is that switch. Calibration procedure for the REM is in the service manual. 


Photios Peter Dalamagas Denova Medical Inc (704)-233-7574