Fri Aug 27 2021
Battery alarm
After changing the battery in the iv pump with a known good battery, the "warning replace battery" message continues to appear and alarm. Is there a way to reset this in a biomed menu or reset the system completely?
   RE: Infusion Pump Report Post
Fri Aug 27 2021Reply from AJ_Diaz
Yes, there should be a button at the bottom of the screen that says something like "change battery" or "reset battery". Should come up as soon as you go into Biomed Mode
Mon Aug 30 2021Reply from Tallahassee CBET
Yes, with the door open turn on the pump. After bootup, with the door still open, press the decimal point key and then the start key.( sometimes you need to toggle back and forth a couple times). you will the see a screen that has date /time, alarm log. below that you will see a reset battery. Ensure you have the unit pugged into AC power before you hit the button.