Connection / Synchronization ( ARGON)
Hi to all,
We have these 2 systems ERBE - APC 300 and ERBE - ICC300, connected together as requested.
Both systems have their own Foot Pedal Switches ( Yellow and Blue ) .
In order for the to work with the ARGON we have to press in parallel the blue foot switch from APC 300 and the blue foot switch from ICC300.
Do you know how we can operate it just doing it only from one pedal ?
Thanks in Advance
RE: Electrosurgical Unit (ESU) Report Post
Sun Sep 05 2021Reply from Periwinkle

Hey,
This is simple and we can assist you with a diagram and with written instructions.
Here is a brief and we can discuss further on emails/call.
1) Take out two wires from Blue pedal of Erbe ICC300 and two wires from Blue pedal of APC300.
2) Take a seperate (third) single pedal foot switch. Connect the two wires from ICC300 in parallel + two wires from APC 300 in parallel. Connect them and you're reday to go.
(On a seperate note, we can provide a single pedal footswitch)
I hope this helps. Happy to discuss further in detail.
Kind Regards,
Periwinkle Kaur
+1 (437) 777-8450 (Whatsapp)
Dr Periwinkle Kaur CMgr
Direcor, ESU Medical Solutions Ltd
www.esumedical.co.uk
sales@esumedical.co.uk
Report Post
Tue Sep 28 2021Reply from pityooka

You will need a connecting cable from APC to ICC, it is a short 20cm cable.4-pin/one side male ,other side female