Tue Aug 31 2021

Patient data

How do I remove patient data from the Konica CS-7 workstation prior to redeployment? Also how to I get to the service log in screen? Thanks

   RE: Picture Archiving and Communication System (PACS) Report Post
Tue Aug 31 2021Reply from MedWrench Crystalyn
avatar placemark

What is the name of the manufacturer and model of the equipment you are working on?


Report Post
Tue Aug 31 2021Reply from daveb medi
avatar placemark

My aplogies.  It's for a Konica CS-7 workstation.  


Report Post
Tue Aug 31 2021Reply from daveb medi
avatar placemark

My aplogies.  It's for a Konica CS-7 workstation.  


Report Post
Tue Aug 31 2021Reply from Sindela
avatar placemark

Hello,

You need to log in as admin. Then you need to go to settings. It should be option to delete images. If you would like , I can help you.

 


Report Post
Tue Aug 31 2021Reply from daveb medi
avatar placemark

Thanks!  Do you know the admin password?


Report Post
Tue Aug 31 2021Reply from Sindela
avatar placemark

It should be in the manual. If you would like, I can share it with you. 


Report Post
Tue Aug 31 2021Reply from Denova Medical
Denova Medical

Default Log in:

User Name: Maintenance

Password: kmmg567

CS-7 can be Windows 7 or Windows 10


Photios Peter Dalamagas Denova Medical Inc (704)-233-7574

Report Post
Thu Oct 21 2021Reply from MASSAMBA
avatar placemark

Le mot de passe Administrateur de CS-7 est  kmmg5678