WATER IN DRAIN

WATER IN DRAIN FAULT. ANY RECOMMENDATIONS? AT END OF EXHAUST STAGE. 733 HC STERILIZER

   RE: Sterilizers/Autoclaves Report Post
Wed Sep 01 2021Reply from Dhakshin
avatar placemark

Check vacuum level during drying phase. It should reach below 1 psi. If it's above 4psi then this alarm will come. Check water pressure, water inlet strainer,  water inlet solenoid valve and finally vantury system. Rarely water lavel detector. 


Report Post
Thu Sep 02 2021Reply from Bob@montaquip
avatar placemark

In general, Water in drain following a drying phase would indicate it is backfilling into drain. Drainline or chamber trap check valve leaking. When unit ceases pulling a vacuum at end of drying phase, advances to air in, water is also being pulled in at the same time. My guess based upon limited information available. If it's actually happening when the machine is exhausting, that is when steam pressure is being evacuated after sterilize, then you have something  a little more complicated going on. 


Report Post
Thu Sep 02 2021Reply from Bob@montaquip
avatar placemark

In general, Water in drain following a drying phase would indicate it is backfilling into drain. Drainline or chamber trap check valve leaking. When unit ceases pulling a vacuum at end of drying phase, advances to air in, water is also being pulled in at the same time. My guess based upon limited information available. If it's actually happening when the machine is exhausting, that is when steam pressure is being evacuated after sterilize, then you have something  a little more complicated going on. 


Report Post
Fri Oct 22 2021Reply from UNM
avatar placemark

LOW WATER PRESSURE, REPLACE WATER VALVE.


Report Post