Hello The device showed ERROR 300 at first and started calibration not knowing I need DOT ligasure for it and I only have Ligasure with QR code, after a restart it shows ERROR 274 that needs calibration. I was wondering if there is a way to bypass it somehow without having dot Ligasure ?
Fri Sep 10 2021Reply from Chris S
Chris S
Unfortunately the dot pattern is required for calibration.
Thu Sep 16 2021Reply from Josh-RenewBiomedical
You can complete the scanner calibration with a made up dot pattern.  Take a piece of paper measuring 2x1 inches.  Make a dot approximately 1/2 inch from the right side.  Make another dot approximately 1 1/2 inches from the right side.  The correct positioning of the dots will allow you to complete the calibration.  Also, dirty scanners can cause a 300 error.  I recommend cleaning them before attempting to complete the calibration.
Josh
Renew Biomedical
Wed Jan 11 2023Reply from Moises L
Please Josh, could you send me a picture about that ? please.
Im  from Ecuador, Thank you
Fri May 26 2023Reply from dutraferreira
Hello Josh!
I'd appreciate if you could send me a picture of the dot pattern needed. I'm going through the same issue... But here the only error presented is E274. So I only need to recalibrate the scanners, but I don't have this dot pattern LigaSure here, only with the QRcode.
Thanks for sharing your knowledge.
Victor.
Wed Jun 21 2023Reply from Josh-RenewBiomedical
Good morning,
I am unable to attach pictures here.  If you do a web search for the ligasure connector, you will see the spacing and image that is needed.  This can be created on a brother printer.  I am also fairly sure a marker and paper/cardstock would provide the same result.
Thanks,