VTBI near end

I have a unit that shows VTBI near end as soon as i start the pump. We test it at 250ml/hr with a set value of 25ml. anyone know what setting need to be change for this notice to go away?

   RE: Infusion Pump Report Post
Mon Sep 13 2021Reply from Ben C
Ben C

One of these companies may be able to assist:

Adepto Medical

Elite Biomedical 

Mobile Medical Repair

iMed


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Mon Sep 13 2021Reply from jpowell
avatar placemark

I'm not sure on what setting, but it is an option for sure. The Medfusion 3500 has the same setting option you can enable or disable. 


Report Post
Mon Sep 13 2021Reply from jpowell
avatar placemark

I'm not sure on what setting, but it is an option for sure. The Medfusion 3500 has the same setting option you can enable or disable. 


Report Post
Tue Sep 14 2021Reply from Adepto Medical
avatar placemark

Hello Jreyes31,

It can be changed using a program called HiBased. This cannot be modified without this program. It is in the modification data tab then under the alarms tab. There is a Pre-alarms VTBI and time setting which is what you would change. If you don't have this program then you can send us your unit and we can do it for you (www.adeptomed.com). Give us a call if you have questions. 

Hope this helps!


Report Post