Mon Sep 13 2021

No probe

When you turn the machine on , on the screen it say no probe . Any ideas what the problem might be?

   RE: Veterinary Equipment Report Post
Mon Sep 13 2021Reply from Ben C
Ben C

What is the name of the manufacturer and model of the equipment you are working on? 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Mon Sep 13 2021Reply from David Yeargin
avatar placemark

SIUI CTS- 200V+


Report Post
Tue Sep 28 2021Reply from David Yeargin
avatar placemark

Does anyone have any ideas, please 


Report Post
Tue Sep 28 2021Reply from T-Lobato
T-Lobato

Hi David,

I don't suppose you have a spare probe...Have you looked it over for damage?  More often than not, they get dropped and it gets damaged inside and out.  If everything looks good, I would reset the system, unplug the probe and then back in.  The unit could just be hung for some reason.  Other than that, if you have a probe leakage tester you could check it that way.  There are a lot of good 3rd party ultrasound companies you could reach out to as well.

Sorry I couldn't be of more help.


TJ Lobato | MedWrench Guru
Report Post
Thu Sep 30 2021Reply from David Yeargin
avatar placemark

Thanks 


Report Post