Failed Boot
RE: Ultrasound Systems Report Post

Philips iU22 E-cart. I performed a front end bypas, and the UMB returned to a normal state (AA), and I got the normal beeps through start up, but I only get a black display. I have swtiched the video cards which did not resolve the issue and pluging in an external monitor is also blank. The iu22 monitor works if plugged into to another iu22, which I did to test if it was working correctly. Any ideas?
Report Post

Replace the power board that sits behind the umb on the back wall. Had the same issue and replacing that took care of it.
Report Post

Boot to black either comes from the video cards or the NAIM. The system clock resides on the NAIM which will cause this issue. If you have a donor system, grab the NAIM and try it. That would explain why you were able to boot with the Frontend Bypass….