Please help me with this one: The #4 ecg pin from the left front of my MX40 is corroded and pitted: what causes this and how can it be corrected?

   RE: Telemetry-Central Station Report Post
Tue Sep 21 2021Reply from Chris S
Chris S

There is an old Philips service bulletin on this issue (SB86202821A).  Philips blames the issue on the cleaning process:

  • Application of disinfecting products other than Isopropyl Alcohol inside the connector on the patient cable can result in corrosion and/or chemical residue build-up
  • Application of product other than Isopropyl Alcohol inside the connector on the patient cable, followed by connection to the MX40/MX4J PWM while there is still fluid inside the patient cable connector, and then insertion of batteries into the device, will cause a reaction between the base metals of the contacts. This then results in corrosion to both the MX40/MX4J connector pins and patient cable connector contacts
  • Application of bleach inside the connector on the patient cable results in corrosion and residue build-up on the contacts.

Philips has recently redisigned the leadset cables to due to a faulty gasket design. Any new trunk cable orders should come with the new design.

As far as a fix, there are not a lot of options.  Attempting to scrape off the corrosion will remove the anti-corrosion coating and will quickly return.  The pack will probably need to be returned to Philips, or a third party for repair.


Official MedWrench Guru
Report Post
Thu Feb 02 2023Reply from willf04034_6646
avatar placemark

We will also be happy to answer any  questions you may have regarding the MX40 service and operator manuals at anytime.


Report Post
Mon Feb 27 2023Reply from mrwedavi
avatar placemark
@Vampire Survivors What is the return process for Philips?

Report Post
Tue Feb 28 2023Reply from Chris S
Chris S
You can contact Philips at (800) 722-9377 and set up an RMA. They will explain the exchange process.

Official MedWrench Guru
Report Post
Tue Sep 26 2023Reply from Euniceda
avatar placemark

Philips has received reports of corrosion of the patient cable connector pins on the MX40/MX4J Patient.


pizza tower 
Report Post
Mon Oct 30 2023Reply from billyroberts
avatar placemark
addressing the corrosion and pitting of the #4 ECG pin on your MX40 involves a combination of cleaning, protection, and, if necessary, replacement. The primary focus should be on preventing further exposure to factors that can cause corrosion and ensuring proper maintenance to keep the device in optimal condition