Error ASP_MSY_E_5003

My C-arm get error: Generator: Reference voltage failure. If failure occurs frequently, inform service. Date time ASP_MSY_E_5003. Anyone have solution for that case. Thank you

   RE: Mobile C-Arm Report Post
Tue Sep 28 2021Reply from Tomelgy
avatar placemark

On the D1 PCB you will find the test point for generator 10+v reference. Check its 10+v.

Also, check the other D1 power suplies are within spec like +/-15v etc.

Does the error appear at: Power-on, standby or when radiation released?

If when radiation releasd, set kV to minimum (manual mode) & see if it appears again.

It could be noise, or X-Ray tube arcing or the D1 PCB faulty.

Tom


Report Post
Tue Sep 28 2021Reply from Duong1711
avatar placemark

Hi Tom.

Thanks for your reply..

my problem occurs when power on and X ray released .Could you share any docs about error codes .

 

BR


Report Post
Thu Oct 14 2021Reply from Duong1711
avatar placemark

Hi Tom.

I already checked, voltage is normall.

In C-arm show some codes:

Err 2070, err 8000.

In PC still recognize main system 

 


Report Post