My C-arm get error:
Generator: Reference voltage failure.
If failure occurs frequently, inform service.
Date time ASP_MSY_E_5003.
Anyone have solution for that case.
Thank you
Tue Sep 28 2021
Error ASP_MSY_E_5003
Tue Sep 28 2021Reply from Tomelgy

On the D1 PCB you will find the test point for generator 10+v reference. Check its 10+v.
Also, check the other D1 power suplies are within spec like +/-15v etc.
Does the error appear at: Power-on, standby or when radiation released?
If when radiation releasd, set kV to minimum (manual mode) & see if it appears again.
It could be noise, or X-Ray tube arcing or the D1 PCB faulty.
Tom
Report Post
Tue Sep 28 2021Reply from Duong1711

Hi Tom.
Thanks for your reply..
my problem occurs when power on and X ray released .Could you share any docs about error codes .
BR
Report Post
Thu Oct 14 2021Reply from Duong1711

Hi Tom.
I already checked, voltage is normall.
In C-arm show some codes:
Err 2070, err 8000.
In PC still recognize main system