Out of nowhere my yag is throwing a 7.3 fault and the water nor cryogen is not heating up. Is there anyway that you can help me or even on the phone and Ill pay for the services as long as I can get this fixed as soon as possible. Thank you very much looking forward to speaking/hearing from you. Im based in Florida.
Wed Sep 29 2021
Fault 7.3
Wed Sep 29 2021Reply from Ben C

Please fill out a REQUEST FORM and we will help find a service company for you!
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Thu Sep 30 2021Reply from Wayne-UML

A 7.3 error deals with the machine not sensing coolant flow. It can be either the pressure switch or a pump related issue. In either case, give our company a call.
Universal Medical Lasers
321-984-3177
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post