Has anyone encounter low flow on these units? I am testing at 50 LPM but only getting about 40 LPM on my tester. I also tested a brand new one out of the box at 50 LPM and it was only slightly better at 42 LPM.
Wed Oct 20 2021
Low flow
Wed Oct 20 2021Reply from Ben C
What is the name of the manufacturer and model of the equipment you are working on?
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Wed Oct 20 2021Reply from awaldron
Wed Oct 20 2021Reply from T-Lobato
Awaldron,
There is a technical manual that can be located on the left under documents. See page 29. By chance are you at high altitude?
Just me.
T.
TJ Lobato | MedWrench Guru
Report Post
Wed Oct 20 2021Reply from awaldron
Thu Oct 21 2021Reply from T-Lobato
The manual also states, "allow the device to sufficiently warm up; rapid changes in temperature can affect the sensor." Oddly, the Acceptance/Peformance Checks on page 14 mention nothing about checking the readings.
TJ Lobato | MedWrench Guru