Battery conditioning test will not finish
I got a brain the user wrote that it failed battery conditioning test(hereby referred to as "BCT") twice. So I do my own BCT and I get this issue. The time to completion is stuck at 00:00 for about 24 hours. So I turn the unit off, check the battery connection pins, bend them slightly to give them better contact with the battery, install a new battery and try again. Same issue, so I leave the brain running for much longer to see if that fixes it with no luck. Lastly, I tried taking apart the unit and putting it back together just in case there was some lose connection causing the problem. No dice. There are no error codes in the logs that has helped me out. So I'm wondering, is this a logic board or a power supply issue, what else should I try and where can I get the correct part?
RE: Infusion Pump Report Post
Wed Oct 20 2021Reply from Ben C

One of these companies may be able to assist:
Find me on Facebook, LinkedIn, YouTube or Twitter
Report Post
Wed Oct 20 2021Reply from Rachel Homier

Aaron, if you'd like to send this into J2S for a free evaluation give us a call or submit an RGA request on our website. We're happy to assist.
844.DIAL.J2S | www.j2smedical.com
Rachel Homier, District Sales Manager
Report Post
Thu Oct 21 2021Reply from rsiefers

I have had similar situations occur. I have found that the switching power supply is the cause. It is located on top of the card cage assembly. Everytime this happened to me, that was the "fix"
I only have an OLD part # (from elite biomedical) is listed as "END OF LIFE"
you may be out of luck, as Carefusion is discontinuing the older style 8015 & 8000's and will only be stocking the newer style pump/parts
Hope this helps