I'm not sure how to perform the Occlusion force 8 lb +- 1 on this unit and I can't view the PM instructions on MedWrench..Do you have instructions for this step in the inspection process?
Wed Oct 20 2021
Occlusion
Thu Oct 21 2021Reply from CTI
What you need to do this test on the Infus O.R. is a Wagner linear force gauge Model FDK20. You then have to position the gauge between the plunger driver and the syringe clamp. Start the pump and when it goes into the occlusion alarm you can read the linear force on the gauge.