Power Supply DDM Samsung H60 Ultrasound

Hi Dear all, Recently i had a problem with the DDM of Samsung H60. What is happening: When i turn ON the power in 220 VAC the machine start up normally, but when i turn ON in 127 VAC, after appears the ECO image in 15 Seconds de power DDM goes shutdown. What was already done: I changed de Filter capacitors from primary side, Checked out all ESR of electrolits capacitors and all are good. PS.: On the table the power it works in 220 VAC and 127 VAC without any problem. Any help it will be appreciate! For while thank you. Cesar Remi

   RE: Ultrasound Systems Report Post
Tue May 10 2022Reply from JohnBad
avatar placemark

Hi, did you solved the problem? I have one but the problem is that is power down after 5 seconds or less...


Report Post
Fri May 13 2022Reply from cesars12
cesars12

Hello JohnBad, this power supply it's working only in 220 VAC. Try to connect then to an 220VAc to check if work fine.


Report Post
Mon Dec 18 2023Reply from HEMSA
avatar placemark

Same issue. System turns off after 5 or 10 seconds.

Before turns off.

A fatal error has occurred. (0xF000005) 


Report Post