Tue Oct 26 2021

The device gives an error E82 E84

Good afternoon. The device gives an error E82 and E84 when changing attachments. If the device is loaded with the attached attachment, then only the ET manipulator works without errors. Recently, the power supply was repaired (the diode exploded and 2 power keys were punched, respectively, the fuse), after repair it worked normally for two weeks. What could be the cause of the malfunction? Thank you for the attention)

   RE: Surgical Laser Report Post
Tue Oct 26 2021Reply from MedWrench Crystalyn
avatar placemark

What is the name of the manufacturer and model of the equipment you are working on? 


Report Post
Tue Oct 26 2021Reply from AlexVidnova
AlexVidnova

I wrote it when creating a thread, but I can't see it. Lumenis Light Sheer Desire


Report Post
Tue Oct 26 2021Reply from AlexVidnova
AlexVidnova

can you help with a diagram for this unit?


Report Post
Tue Oct 26 2021Reply from AlexVidnova
AlexVidnova

burnt out 40N25 keys were replaced with IRF4242, only one pair. Maybe that's the problem, although it worked fine for 2 weeks. With a pulse, both capacitors are used, maybe one is not pumped to the desired values?


Report Post