Error 7300 and 7310

I have a C arm they changed a card now an error appears7300

   RE: Mobile C-Arm Report Post
Thu Oct 28 2021Reply from Tomelgy
avatar placemark

Assuming errors with: MSY_7300 & MSY_7310 you have a problem with X-Iris limits.

What baod was replaced? Was it CPU D1? Were taken before board replaced & were Backups restored after board replaced?

Does the collimator run during power-up sequence. Have you checked connectors on collimator?

Has a plug been left out?

Tom


Report Post
Fri Oct 29 2021Reply from Bioteck
avatar placemark

the electronic card that was changed is the serial number CIPP 80 89 555 D66 E1 It is the one with the bnc video input connector.


Report Post
Fri Oct 29 2021Reply from Tomelgy
avatar placemark

The D66 CIPP board has nothing to do with collimator function.

Maybe you are just unlucky, You need to investigate the problem further.

Does the collimator run the search run during the power-up phase.

Have you checked the SG cable & connector pins at X10

Tom