i have mythic 18 hematology analyzer, it was ok but since last week it shows startup cycle failed and QC not done...plz help me out
Sat Oct 30 2021
Startup & QC
Tue Nov 30 2021Reply from Paul Moyer
Mon Jan 31 2022Reply from Gile

Try bleach cycle,if not work ok afther that, dismantling needle rinsing head and clean it in bleach and replace o-ring.
Report Post
Mon Jan 31 2022Reply from Gile

Try bleach cycle,if not work ok afther that, dismantling needle rinsing head and clean it in bleach and replace o-ring.
Report Post
Tue Feb 01 2022Reply from ismaeel09

Thank you Paul Moyer & Gile. There was a problem with RBC counting chamber, replaced. Can RBC counting chamber be repaired as i have old one available? Apparently there is no breakage or damage
Report Post
Tue Feb 01 2022Reply from Gile

In a small number of cases, I solved the problem myself by crimping the connectors again. In most cases, that didn't help
Report Post
Tue Feb 01 2022Reply from ismaeel09