Steris ETO abator not coming up to temperature

The abator for my 3017 Steris ETO sterilizer is not reaching temp, its stuck at room temp rather than the typical ~300F. Im looking for any ideas as to why. I have two identical units and neither will heat up. This just started occurring as it worked fine a couple of days ago

   RE: Sterilizers/Autoclaves Report Post
Tue Nov 02 2021Reply from Ray Brown, CBET
avatar placemark

We don't have these devices here, but if you have two of them side-by side, I'd suspect the power wiring coming in, especially if they run on 240v. It's possible that something's happened to the wiring or the circuit breaker of L2, the second leg, while the electronics run on L1. So I'd check all the fuses and reset the circuit breakers and see if that clears it up.