Setting correct time for history

How to imput correct time when searching history

   RE: Vital Signs/Multi-Parameter Report Post
Mon Nov 08 2021Reply from MedWrench Crystalyn
avatar placemark

What is the name of the manufacturer and model of the equipment you are working on? 


Report Post
Mon Nov 08 2021Reply from Elquadero
avatar placemark

GE farscape v100 dinamap


Report Post
Mon Nov 08 2021Reply from Elquadero
avatar placemark

Sorry farscape!


Report Post
Mon Nov 08 2021Reply from Elquadero
avatar placemark

Auto correct frustrates me...

CARESCAPE!


Report Post
Tue Nov 09 2021Reply from Chris S
Chris S

Enter configuration mode:

With the monitor off, press and hold the Menu button at the same time as pressing the On/Off button until the display test completes. CFG displays in the Systolic window. As the monitor turns on in the configuration mode, a brief display appears showing the software revision, NIBP technology, and temperature technology of the monitor. These displays appear only during the first part of the power up sequence and are not selectable and cannot be changed.

1. Press the Menu button to move from one setting to another. Use the +/- buttons to increment or decrement the setting.

2. To exit the configuration mode, press the On/Off button.

3. To continue with other changes, press the Menu button. CFG appears in the Systolic window. To change parameter settings, press the Menu button and select the parameter

NOTE

For the date and time to be saved, you must advance the menu through the minute setting.


Official MedWrench Guru
Report Post