Sabrena,
Error 1530 is related to a mismatch between 2 energy detectors. The reason the issue pops up intermittently is probably because it is on the threshold of the error. It is doubtful that there is an optical or electrical issue, but that the laser will need to be calibrated.
I hope this helps.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post
Hi everyone.
I'm looking for a Service Manual or Calibration Manual for Cutera Enlighten Laser.
Thanks in advance
Report Post

The 1126 error doesn’t appear in the manuals, but recalibration seems to take care of it. Also, I have found that the LCR is starting to lose integrity which means 1125 may start popping up once the 1126 is cleared.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post
Hello, Im having multiple errors with the enlighten laser 2, is aanyone able to please send me a service manual, Im so appreciative. katy@abint.com.au
Report Post
@Wayne-UML how did you recalibrate the laser to clear the 1126 error? Is the recalibration in a manual somewhere?
Report Post

Yes, it’s a process found in the manual. A good amount of our business is in Enlighten repair. We can come to you if you need something.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post