Change the language

I would like to change the Philips - ATL HDI 5000 ultrasound machine to French language. Initialy, it is German. Thanks..

   RE: Ultrasound Systems Report Post
Fri Nov 12 2021Reply from MarioLM
avatar placemark

Hello, try this to get into service mode. Maybe there you will have the language options (I don't remember if that is where it was, it's been a long time since I worked on a HDI 5000).

Essaye ça

To get to service mode: Press SUPERKEY & 0 you will be in Engineer panel

A pop up window asking for USER: Type CSR and return

A pop up window asking Password: type USER and return, select OK


Report Post
Fri Nov 12 2021Reply from nbaever3
avatar placemark

thank you for sharing your reply with us, this is really helpful for me


Report Post
Wed Dec 01 2021Reply from nbaever3
Wed Dec 01 2021Reply from dferguson
dferguson

If I remember right, the language is tied to the licensed option file.  


 
 

Report Post
Sun May 08 2022Reply from talyanya
avatar placemark

To the community. I will explain how to change the options, and turn on the ones needed on the Philips ATL HDI 5000 ultrasonic machine, maybe someone will then help me get a service key for the 512 redwood.

To change the options you need to have a magneto-optical disk. Insert it into the device and format it. Then press the superkey and the “0” key and the C text at the same time.

The time on the device will stop, 10-15 minutes and the device will write the information to the magneto-optical disk, and throw it out about.

The MO disk must be inserted into external media, such as a computer, and opened. Copy file MACHOPTS.DAT to your computer.

Open this file with notepad ++

You see functions and options with changes 0 and 1.

Respectively 1- works, 0-no

The number of options working in the device is constant. But you need to decide what you need to change - language, sensor, cardio and the like.

In order to enable what you want, you need to disable what you consider unnecessary.

To do this, right-click in the program notepad ++

Perform the action - cut, paste.

https://drive.google.com/file/d/15Nh8VchnaNWwB9q7JRBCnES1Yh-Xz_6r/view?usp=sharing

To the comma, before 0 or 1, cut out the whole inscription, and rearrange it to the desired option, and the one that was there to cut and paste in place of the previous one.

Save changes.

Format MO drive. Create the PEG_HD folder on the computer, the SITE folder in it, the MACHCFG folder in it, and insert the edited MACHOPTS.DAT file there.

 Burn this folder to the MO drive.

 Then insert the MO disk into the ultrasonic machine, press the superkey and key "0", enter the login "csr" and password "user" select options, and read new options, confirm actions, and reboot the device - and you need the options included.


Report Post
Sun May 08 2022Reply from talyanya
avatar placemark

To the community. I will explain how to change the options, and turn on the ones needed on the Philips ATL HDI 5000 ultrasonic machine, maybe someone will then help me get a service key for the 512 redwood.

To change the options you need to have a magneto-optical disk. Insert it into the device and format it. Then press the superkey and the “0” key and the C text at the same time.

The time on the device will stop, 10-15 minutes and the device will write the information to the magneto-optical disk, and throw it out about.

The MO disk must be inserted into external media, such as a computer, and opened. Copy file MACHOPTS.DAT to your computer.

Open this file with notepad ++

 

You see functions and options with changes 0 and 1.

Respectively 1- works, 0-no

The number of options working in the device is constant. But you need to decide what you need to change - language, sensor, cardio and the like.

In order to enable what you want, you need to disable what you consider unnecessary.

To do this, right-click in the program notepad ++

Perform the action - cut, paste.

https://drive.google.com/file/d/15Nh8VchnaNWwB9q7JRBCnES1Yh-Xz_6r/view?usp=sharing

To the comma, before 0 or 1, cut out the whole inscription, and rearrange it to the desired option, and the one that was there to cut and paste in place of the previous one.

Save changes.

Format MO drive. Create the PEG_HD folder on the computer, the SITE folder in it, the MACHCFG folder in it, and insert the edited MACHOPTS.DAT file there.

 Burn this folder to the MO drive.

 Then insert the MO disk into the ultrasonic machine, press the superkey and key "0", enter the login "csr" and password "user" select options, and read new options, confirm actions, and reboot the device - and you need the options included.


Report Post
Sun May 08 2022Reply from talyanya
avatar placemark

To the community. I will explain how to change the options, and turn on the ones needed on the Philips ATL HDI 5000 ultrasonic machine, maybe someone will then help me get a service key for the sequoia 512.

To change the options you need to have a magneto-optical disk. Insert it into the device and format it. Then press the superkey and the “0” key and the C text at the same time.

The time on the device will stop, 10-15 minutes and the device will write the information to the magneto-optical disk, and throw it out about.

The MO disk must be inserted into external media, such as a computer, and opened. Copy file MACHOPTS.DAT to your computer.

Open this file with notepad ++

 

You see functions and options with changes 0 and 1.

Respectively 1- works, 0-no

The number of options working in the device is constant. But you need to decide what you need to change - language, sensor, cardio and the like.

In order to enable what you want, you need to disable what you consider unnecessary.

To do this, right-click in the program notepad ++

Perform the action - cut, paste.

https://drive.google.com/file/d/15Nh8VchnaNWwB9q7JRBCnES1Yh-Xz_6r/view?usp=sharing

To the comma, before 0 or 1, cut out the whole inscription, and rearrange it to the desired option, and the one that was there to cut and paste in place of the previous one.

Save changes.

Format MO drive. Create the PEG_HD folder on the computer, the SITE folder in it, the MACHCFG folder in it, and insert the edited MACHOPTS.DAT file there.

 Burn this folder to the MO drive.

 Then insert the MO disk into the ultrasonic machine, press the superkey and key "0", enter the login "csr" and password "user" select options, and read new options, confirm actions, and reboot the device - and you need the options included.


Report Post