ERROR F56 AND NOT RESPONDING TO ADJUSTMENT MODE
look up F56 in the manuals, also, enter adjustment mode and adjust the (+) and (-) side of the battery balance.
Manual can be found HERE but MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument, and the MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges.
hello, i have same problem,
the battery voltage on the multimeter is -120v, when I set this value in the machine after switching off and on again the value goes to -145v, the measured value is not memorized.
what board can i replace to fix the problem?
thanks.
hello, i have same problem, by decreasing the number of batteries on the + or - side, I managed to enter the battery voltage configuration, but there is the (-) side, the measurement channel is not stable.
thanks
Hi the problem is always the battery the voltage on the + side most not be high than the - side +120 and -120 both side any deviation will trigger error f56 .