starlux 1540
I recently had the same issues and after looking into it discovered the calibration is actually just a power check so if you have a power meter and have the service manual I have just been doing power check myself.
Sean Casey
BMET III
Sodexo Clinical Engineering
SeanC is correct. The calibration process is the laser making sure it has sufficient power and making voltage adjustments to ensure proper output. If it is failing, it could be several things. The calibration fiber, which is part of the handpiece, could be bad; the laser rod may need replacement; or some optics or the main fiber may be burnt.
Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Wayne Delgado
www.umlasers.com
Thanks for your input Wayne, I actually took the laser repair class with Greg Abstein and yourself last September. Hope you are doing well.
Sean Casey
BMET III
I have problems with lux g handpiece. Just replaced lamp and need to set calibration parameters. Does anyone know how to set it?