OEC - How to Reset Image Rotation?

I have a question regarding operating the OEC 9800/9900 C-arms. Is there a way to "Reset to Home" the image orientation using the control panel? We can rotate the image digitally CW or CCW using the control panel. But is there a way to reset to default (0 rotation) through the control panel or menus of the software? Resetting the C-arm resets the image orientation, but of course, we cannot do that in the middle of a case.

   RE: Mobile C-Arm Report Post
Mon Nov 29 2021Reply from C12Guy
C12Guy

No that I know of, Just the Home position in the operators menu.


Report Post
Mon Nov 29 2021Reply from Sam Aman
avatar placemark

Would you please comment on how I can get to the operator's menu? Is that possible after I start taking fluoro? Thanks.


Report Post
Mon Nov 29 2021Reply from C12Guy
C12Guy

To access Setup features, press the Setup key on the Workstation keyboard.

Select C-arm last button on the right side,

Set defaults
To set the default position for camera rotation, collimator rotation, collimator iris and leaf position and magnification:
1. Select Home or Retain Last for the default position.
Home returns the camera and collimator to the home position, opens the collimator iris and/or leaves to a fully open position, and returns magnification to normal.
Retain Last keeps the same settings for the next patient exam or at start-up.
2. Select OK to close the C-arm screen and return to the Setup screen.
3. Select Exit to close the Setup screen.

 

Key there is the New Exam so it will not change in the middle of an exam.


Report Post
Mon Mar 14 2022Reply from Sam Aman
avatar placemark

Perfect. Thanks.


Report Post
Mon Mar 14 2022Reply from Sam Aman
avatar placemark

Perfect. Thanks.


Report Post
Mon Mar 14 2022Reply from Sam Aman
avatar placemark

Perfect. Thanks.


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree