Many IV pumps with same error

Clean Load port # 2. I have many pumps that go down with this error. Any suggestions. Thank you Andy

   RE: Infusion Pump Report Post
Wed Dec 01 2021Reply from Chris S
Chris S

Baxter's suggestion is to clean load point 2 with a cotton swab and distilled water.  This doesn't seem to work very often.   

The part that is usually replaced when I send these pumps to Baxter or 3rd party companies is the ultrasonic sensor (Part number 35711-S-988).  I haven't been able to find this part for sale anywhere.


Official MedWrench Guru
Report Post
Thu Jun 09 2022Reply from Matt O
avatar placemark

Depending on the software version of your pumps there is also a calibration procedure that I have had some success with.  It requires a different service passcode.  You'll know if your software will allow you to calibrate if the pump takes the password.  If it doesn't, your software is not the right version and you'll most likely have to send the pump in for repair.  The other password is 8961.


Report Post
Mon Jul 17 2023Reply from hartfighting
avatar placemark

A separate service PIN is needed for this.  Pump acceptance of the password will indicate whether or not your program supports calibration. It doesn't appear like this is very effective.   


drift boss
Report Post
Mon Jul 17 2023Reply from amusedthrushe
avatar placemark

This requires a different service PIN.  Whether your program enables calibration will be shown by the pump's acceptance of the password. This doesn't seem to be working all that well.   