Thu Dec 02 2021

Error MD10024

Console Errol [ MD10024 ] An error notification was issued by the Imager Reader . The target image may have been input erroneously in the study of details below : Patient ID : Patient name : Study date : 2021/12/02 Study time : 11 : 22 : 38.284 > The image is registered to the study of details below : Exposure menu : PELVIS IP number : a40000008c + Perform re - exposure if necessary . Then , subject the IP to image erasure processing on the Image Reader . OK RU : IP being read

   RE: Computed Radiography Report Post
Thu Dec 02 2021Reply from Luis_Med
avatar placemark

Hello,

I recommend you go re install console. I can share the service manual with you.


Report Post
Mon Dec 06 2021Reply from MedWrench Crystalyn
avatar placemark

If you would like to share with the MedWrench community, please click HERE and we will upload it to the product page on your behalf.


Report Post
Sat May 21 2022Reply from wcs007_4905
avatar placemark

corrupted software or database, if you use worklist there may be incorrect data in the registration


Report Post