Error 7150: Slot disphragm index point missed.
Check the collimator & cables. Check the nylon gear teath for damage. Make sure the slot diaphragm moves during boot-up.
Check the 'lift solenide voltage' when its activated during boot cycle.
Error 9122, was this error 9102: Slot diaphragm lift solenoide failure.
Error 4028: Acqusition aborted.
Check faults on the Memoskop acquisition cable (D1.X10.27).
Maybe this is related to E7150 but not normally if E7150 can be cleared after boot-up.
Report Post
E7150: Usually this is caused by stripped teeth on the lower Slot Diaphragm ring gear
E9122: (E9122 not found) If this was actually E9102, then a bad Slot diaphragm lift solenoid is probably the cause (they are rated for intermittent operation only, and can overheat—look for a blackened solenoid). This may also be the cause of the E7150
E4028: many times an intermittent E4028 can be caused by a worn socket on the SG Cable connector, but it also can be a bad Memoskop failing during operation.
Let me know if you need replacement parts.
- Wayne H.
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Errors that start with an "E" are "non-fatal" and can usually be reset by pressing any switch, including the fluoro pedal. Errors that start with an "F" are "fatal" errors, and usually require a reboot.
The SIREMOBIL Error classes are:
· I = Informations
· W = Warnings
· E = Errors
· F = Fatal Errors
- Wayne H.