Tue Dec 21 2021

Fluent System PM Check or Service Manual

Looking for information on the PM procedure for the Fluent system. We have one that is not used that often and Hologic is wanting 5K for service.. I believe it is recommended every 2 years, but with Covid shut downs and the infrequency of use I am looking for other options.

   RE: Fluid Management Systems Report Post
Tue Mar 15 2022Reply from SHBMET
avatar placemark

The Fluent Myosure Operators Manual has a "Annual Inspection and Testing" section within it. It is a very easy PM to do and as long as the two tests check out you should not have to pay for any service.

Load Cell Calibration Check

Pressure Calibration Check

The Operators Manual is available through Hologic online if you do a google search for:

"Fluent Fluid Management System Operators Manual"


Report Post
Tue Mar 29 2022Reply from MedWrench Manager
MedWrench Manager

Operation Manual can be found here.


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.